Hi All,
I just bought a DMM6500. I tried to measure temperature with a type K thermocouple. I let the instrument warm up for an hour & then tried the test. The meter is showing a temperature about 15 degrees C higher than what is correct (at ambient of about 12 degrees C). This can be adjusted in the settings, but if you turn off the instrument & turn it back on, the same error occurs. This is using the front panel connections. The same thing happens when using the rear connections. Other measurements such as current & voltage, diode, capacitance seem to be good but I have not done an intense comparison with 2 other instruments which do match, being a Brymen 629 & a Siglent 4.5 digit bench meter. Although I bought this brand new from the New Zealand Tektronix agent Vicom on Thursday last week, the unit is about a year & a half old according to the calibration sheet, which does not impress me much, being a retired calibration technician.
So my question is: Is this unit faulty, or does it need re calibration, or am I not doing the test correctly, or does the reading have to be calibrated every time you turn on the instrument ?
Any help would be appreciated.
Thanks,
Gary
Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.
Temperature measurement big discrepancy
-
- Keithley Applications
- Posts: 1514
- Joined: October 15th, 2010, 10:35 am
- Country: United States
- Contact:
Re: Temperature measurement big discrepancy
If it is 1.5 yrs old, I’d first check the firmware. Version 1.7.2 is current release.
For the thermocouple on front or rear terminals, you are providing a simulated reference junction temperature. That would be an estimate of the temp at those terminals or the room temp. I’d expect a room temp closer to 20 deg C rather than 12.
Replace the T/C with a shorting jumper HI to LO (ideally cooper wire). Does the meter now report the temperature of your simulated reference junction?
For the thermocouple on front or rear terminals, you are providing a simulated reference junction temperature. That would be an estimate of the temp at those terminals or the room temp. I’d expect a room temp closer to 20 deg C rather than 12.
Replace the T/C with a shorting jumper HI to LO (ideally cooper wire). Does the meter now report the temperature of your simulated reference junction?
Re: Temperature measurement big discrepancy
Hi Andrea,
Thanks for the info. I did the upgrade. There is a definite improvement. It still takes about 15 minutes for the temperature to stabilize. Is this correct ?
Best regards,
Gary
Thanks for the info. I did the upgrade. There is a definite improvement. It still takes about 15 minutes for the temperature to stabilize. Is this correct ?
Best regards,
Gary
Re: Temperature measurement big discrepancy
Hi Andrea,
After waiting for the temperature to stabilize & with 2 of the same type thermocouples placed next to each other, one plugged into the DMM6500 & the other into a Brymen BM869s meter, I adjusted the temperature reading. After this I placed a short on the i/p terminals of the DMM6500. The reading is 22.6 C, which I presume is correct.
Thanks for your help. I have more questions, I will post a separate topic.
Gary
After waiting for the temperature to stabilize & with 2 of the same type thermocouples placed next to each other, one plugged into the DMM6500 & the other into a Brymen BM869s meter, I adjusted the temperature reading. After this I placed a short on the i/p terminals of the DMM6500. The reading is 22.6 C, which I presume is correct.
Thanks for your help. I have more questions, I will post a separate topic.
Gary