synchronize NPLC around zero of 50Hz power line

Post Reply
CH2009
Posts: 8
Joined: October 13th, 2015, 11:01 am
Country: France

synchronize NPLC around zero of 50Hz power line

Post by CH2009 » November 27th, 2015, 9:29 am

Hello,

I have a question about NPLC configuration for a dmm 2701 using a multiplexing card 7706. I would like to know if it is possible to perform a measure with NPLC inferior to 1 and centered around zero of 50Hz power line?

Thanks in advance

Norbert W
Keithley Applications
Keithley Applications
Posts: 227
Joined: October 21st, 2010, 7:18 am
Country: Germany

Re: synchronize NPLC around zero of 50Hz power line

Post by Norbert W » November 30th, 2015, 1:24 am

Hello,

I assume you are going to use your DMM in DCV mode.
Here you can adjust the NPLC setting as you think of appropriate, e.g. select 0.1 NPLC if a fast
measurement is recommended and higher noise can be tolerated &c..
Generally the 2701 A/D converter is of an integrating type, i.e. the best 50 Hz suppression will
be with an integer number setting for the NPLC parameter.