Hello there
A newbie question: We are measuring electrical potentials in plants with the Keithley Multimeter 2701, using 4 channels of the 7700 switching module. We run our measurements with the KickStart software and I have problems with setting up the correct sampling/display rate: There seems to be only a minimum interval of 1 sec (delay), but we'd need a much faster display of the spikes we would like to record/measure. Is there a way to set this up correctly within the software or the machine?
Thanks for your help.
Re: 2701/7700 with HiLink - sampling rate issue
Post by Dale C » June 8th, 2016, 6:00 am
In Kickstart , in the Channel Configuration sheet, Select the RATE as "Custom" then set NPLC as 0.01.
Select RANGE to be fixed (example 10V).
Also, disable the filter, this default but it is worth mentioning.
Now in the SCAN LIST sheet, disable the "Auto Delay, then select Delay to 0.0
These few parameter setting should speed up the process quite a bit.
Hope this helps.
marcusmaeder
Country: Syria
Post by marcusmaeder » June 14th, 2016, 12:33 am
Thank you so much, that helped a lot!
marcusmaeder
Country: Syria
Post by marcusmaeder » June 23rd, 2016, 5:50 am
I saw that there's an issue with the NPLC value: If I set it to 0.01, KickStart says: NPLC value is outside of the allowable range: 0.05 to 50.
There is no way to increase the NPLC to, let's say 0.001 ?
Thank you.
Post by Dale C » June 23rd, 2016, 7:23 am
You are correct. Not sure why it limits t0 0.05 in Kickstart.
On the 2701 you can set it to 0.01, but it is Kickstart that limits this.