Hi.
using the 2701 + 7700 and KickStart, when selecting the CONT function an unhandled exception error. please advise
Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.
2701 Cont function - KickStart
-
- Keithley Applications
- Posts: 51
- Joined: September 17th, 2010, 7:55 am
- Country: United States
Re: 2701 Cont function - KickStart
Hi Zvikaef,
I apologize for the tardy reply, The CONT function on the 2701 in KickStart is a known bug. However, the CONT function was primarily designed for front panel use with the 2701 as it indicates via audible sound ("beep") if continuity is detected. As a workaround, you may instead use the 2W resistance function to gather resistance readings on your devices and you may set limits on the channel for resistances that you would consider as "short" and "open".
Regards,
Jennifer C
I apologize for the tardy reply, The CONT function on the 2701 in KickStart is a known bug. However, the CONT function was primarily designed for front panel use with the 2701 as it indicates via audible sound ("beep") if continuity is detected. As a workaround, you may instead use the 2W resistance function to gather resistance readings on your devices and you may set limits on the channel for resistances that you would consider as "short" and "open".
Regards,
Jennifer C