Open / Close Socket locks up 3706A
System 3706A with 1.53c, has (2) 2602A SMU TSP linked to it. Use a standard TCP/IP socket connection (port 5025) to open and close a socket over and over again (thousands of times), on 3706A, and although it will allow the connection and simple display.settext("hello") type commands, not much else will respond.
One clue, if you press the "EXIT (LOCAL)" button on the 3706A front panel, it will start working again on the next socket connect.
Tried every reset function I can find. Tried abort functions. Tried display.sendkey(75). local lock on and off. display.getlastkey() doesn't return anything.
Obviously, there is something running on one of the three devices, that's the 3706A is waiting for, and the EXIT button aborts that function.
Most of these systems will run for days or weeks without hanging up, some hang up all the time. It's killing production up time.
Re: Open / Close Socket locks up 3706A
"Fatal Error e02 p2D" "Disconnect DUT and cycle power" ????
Can you shift to use of the VXI-11 connection rather than raw socket?
TCPIP0::XXX.XXX.XXX.XXX::inst0::INSTR
int32_t simple_loop_test( void )
struct sockaddr_in dest;
const char *ipaddress_default = { "192.168.1.235" };
int port_default = 5025;
int mytrue = 1;
int sockfd = 0;
int bytes_read;
int bytes_read_cumm;
char rtn_buffer[MAXBUF];
char send_buffer[100];
int loopcount = 1;
while( 1 )
printf( "LOOP%i:\n" , loopcount++ );
sockfd = socket( AF_INET, SOCK_STREAM, IPPROTO_TCP );
if( sockfd <= 0 )
perror("Socket");
return false;
//---Initialize server address/port struct---//
printf( "Connecting to SMU @ IP Address: %s\n", ipaddress_default );
bzero(&dest, sizeof(dest));
dest.sin_family = AF_INET;
dest.sin_addr.s_addr = inet_addr( ipaddress_default );
dest.sin_port = htons( port_default );
//---Connect to server---//
if( connect(sockfd, (struct sockaddr *)&dest, sizeof(dest)) != 0 )
perror("Connect");
return false;
if( setsockopt( sockfd, IPPROTO_TCP, TCP_NODELAY, &mytrue, sizeof(mytrue) ) != 0 )
perror("setsockopt");
return false;
if( ioctl(sockfd, FIONBIO, &mytrue) != 0 )
perror( "ioctl" );
return false;
strcpy( send_buffer, "reset()\n*IDN?\n" );
write( sockfd, send_buffer, strlen(send_buffer) );
bytes_read_cumm = 0;
while( 1 )
memset( rtn_buffer, 0, sizeof(rtn_buffer) );
bytes_read = recv( sockfd, &rtn_buffer[0], sizeof(rtn_buffer), 0 );
if( bytes_read > 0 )
printf( "Recv (%i bytes): %s", bytes_read, rtn_buffer );
bytes_read_cumm += bytes_read;
if( bytes_read_cumm == 54 )
break;
printf( "All 54 bytes expected, where received.\n" );
strcpy( send_buffer, "print(node[2].description)\r\n" );
printf( "SENT: %s", send_buffer );
while( 1 )
if( bytes_read_cumm == 41 )
break;
printf( "All 41 bytes expected, where received.\n" );
strcpy( send_buffer, "print(node[3].description)\n" );
while( 1 )
break;
if( sockfd == 0 )
return( 0 );
shutdown( sockfd, SHUT_WR );
usleep( 100000 );
while( 1 )
bytes_read = recv( sockfd, rtn_buffer, 1, 0 );
if( (bytes_read == 0) || (bytes_read == -1) )
break;
close(sockfd);
return( true );
OUTUPUT RESULTS:
LOOP STARTED @0
Connecting to SMU @ IP Address: 192.168.1.235
Recv (54 bytes): KEITHLEY INSTRUMENTS INC.,MODEL 3706A,04081452,01.53c
SENT: print(node[2].description)
Recv (41 bytes): Keithley Instruments SMU 2602A - 1314283
SENT: print(node[3].description)
Recv (41 bytes): Keithley Instruments SMU 2602A - 1291771
LOOP2695:
LOOP2696:
LOOP2697:
^C
# ./smu --ham
-executing: "./smu --ham"
LOOP1:
int mytrue = 1;
int sockfd = 0;
int bytes_read;
int loopcount = 1;
while( 1 )
if( sockfd <= 0 )
perror("Socket");
return false;
perror("Connect");
return false;
return false;
perror( "ioctl" );
return false;
while( 1 )
break;
while( 1 )
break;
while( 1 )
break;
if( sockfd == 0 )
return( 0 );
usleep( 100000 );
while( 1 )
break;
close(sockfd);
return( true );
OUTUPUT RESULTS:
LOOP STARTED @0
LOOP2695:
LOOP2696:
LOOP2697:
^C
# ./smu --ham
LOOP1:
I'll do some testing on my end.
If feasible, can you see if you experience the issue with only the 3706A (and not the TSP-LINKed instruments)?
With your comment about pressing exit button and it will work on next attempt to connect leads me to believe you have a dead socket. If you can detect this condition, your error handling could try connecting on port 5030 and then close, then retry on port 5025.
Yep, I pulled the relay cards from the 3706A, and disconnected the TSP link completely/physically, removed all node[2] and node[3] command calls, and the error still happens. Tried calling 5030 port, it made no difference. The socket isn't completely dead, since I can still make display.settext() calls and see the display update.
V. Woerdeman also contacted me, and is trying this for me right now.
Wade
Wade