A combined continuity and IR system for testing multiple conductors using a four-wire DMM
xyz6692
Post by xyz6692 » March 1st, 2016, 11:08 pm
Hi All,
I have a 3706A with four 3720-ST cards. I want to use them to measure the continuity and Insulation Resistance of a multi-connector cable as shown in Fig. 7-7 of Keithley Switching Handbook (see attachment). I can't find any channel configuration example in either the user manual or reference manual.
Could anyone help?
Thanks in advance.
F
Screen Shot 2016-03-01 at 12.36.02 PM.png (193.99 KiB) Viewed 11537 times
brian.d.smith
Re: A combined continuity and IR system for testing multiple conductors using a four-wire DMM
Post by brian.d.smith » March 2nd, 2016, 4:10 am
The 3700A Reference Manual page 2-96 begins the discussion concerning "Pole settings and channel patterns". 4-wire is called 4-Pole.
Page 7-16 of the Switching Handbook shows the connections for source and sense. Channels 1-30 are source. Channels 31-60 are sense.
xyz6692
Post by xyz6692 » March 7th, 2016, 11:27 pm
Hi Brian,
Thanks for your reply.
Unfortunately, your reply is for a conventional 4-wire measurement as shown in the Fig. 7-11 in page 7-16 of the Switching Handbook. What I wanted to measure was illustrated in the graph of my original post, the Fig. 7-7 in page 7-11 of the Switching Handbook.
Dale Cigoy from Keithley pointed out to me that the only way to configure as in Fig. 7-7 was to hardwire directly from the Dsub-15 to the outputs of the installed card. For example,
Connect
Dsub-15.1 (2W-Hi) to Slot.3.Output.Lo
Dsub-15.2 (2W-Lo) to Slot.1.Output.Lo
Dsub-15.3 (4W-Hi) to Slot.3.Output.Hi
Dsub-15.4 (4W-Lo) to Slot.1.Output.Hi
Closing slot.1.1015 and slot.3.1015, I can 4-wire-measure a resistor in the exactly same way as in Fig. 7-7 of the Switching Handbook.
Why doesn't Keithley have one pole setting for the backplane A1 and A2, so I can close 1911.Lo to slot.1.output.Lo and 1912.Lo to slot.1.output.Hi, and the same for slot.3?
F
Kenneth P
Post by Kenneth P » March 9th, 2016, 9:11 am
The analog backplane switches are single pull, double throw type. Closing any of the backplane switches close both Hi and Lo at the same time. The Hi and Lo lines cannot be independently closed.
A 3732 Matrix card could give you more flexibility with line routing to the analog backplanes however I believe you don't have such a card.