Can i connect 2300 power supplies in series?
ctraines
Post by ctraines » August 8th, 2014, 7:37 am
I need a 22V battery simulator. Can I put 2 2300 power supplies in series?
Re: Can i connect 2300 power supplies in series?
Post by Dale C » August 8th, 2014, 1:49 pm
Per our spec, for this whole series (you will find the spec under General), the isolation (Low-Earth) is 22V. So you could series two units but max voltage cannot exceed 22V. If you exceed that, damage to the supplies could result.
sanimirza
Post by sanimirza » December 22nd, 2014, 3:26 am
Hi Forum,
I have a 2303-PJ with firmware A03/B01, while calibrating it turned out the manual was not telling the proper things, calibration for the 500mA range I only could to by using the 4 Ohm resistor and was not possible with the 30 Ohm as was stated in the manual..
Post by Dale C » December 22nd, 2014, 11:52 am
Checking with service on this issue.
I will post when it is answered.
Post by Dale C » December 22nd, 2014, 12:11 pm
4ohm for 500ma adjust
The manual is wrong..
Return to “Series 2300”
