Keithley 2300 Series Stability
BrianJCongdon
Post by BrianJCongdon » January 12th, 2015, 9:20 am
I am looking for a supply with extremely high stability/low drift. I need to source up to 1A and will never exceed 24V (I could go as low as 5V.) What do you suggest?
Re: Keithley 2300 Series Stability
Post by Dale C » January 12th, 2015, 10:56 am
I would suggest Model 2200-30-5.
This is a single channel power supply that can output up to 30V at 5A.
The specs are attached.
Series2200.pdf
(295.92 KiB) Downloaded 1047 times
BrianJCongdon
Post by BrianJCongdon » January 12th, 2015, 1:38 pm
I searched the 2200 series document and couldn't find a Stability or Drift spec. The data sheet for the 2300 series lists Stability as 0.01% + 0.5mV over an 8 hour period. Ideally, I'm looking for a supply that is more stable than the 2300 series; anything 5V or above will work.
Post by Dale C » January 12th, 2015, 1:46 pm
The Model 2300 series seems to be the only supplies that have stability and drift specs.
Even the SMUs(source measure units) do not have these specs.
BrianJCongdon
Post by BrianJCongdon » January 14th, 2015, 3:47 pm
Can you provide me with the Ripple and Noise specs for the 2300 series? The Specifications section of the downloadable instruction manual (Appendix A) is blank.
Post by Dale C » January 15th, 2015, 5:50 am
SPEC-2308 (A - Jul 2008)(Customer).pdf
(143.83 KiB) Downloaded 1069 times
2306RevA_DocSpec.pdf
(53.32 KiB) Downloaded 1065 times
2304AspecsrevA.pdf
(125.16 KiB) Downloaded 1046 times
2302RevA_DocSpec.pdf
(49.28 KiB) Downloaded 1046 times
BrianJCongdon
Post by BrianJCongdon » January 16th, 2015, 7:28 am
I'm specifically looking for the noise and ripple specs for DC Voltage output (Battery.) The only reference I can find to noise in the documentation is in the Pulse Current Measurement Operation section.
Post by Andrea C » January 16th, 2015, 8:59 am
The 2300 Series make use of a switching power supply to generate their output voltages and make use of an integrating A/D converter for the measurements.
I suggest you again consider one of the linear power supply products such as the 2200-32-3.
I've uploaded an application note on understanding the Linear Power Supply Specs including concepts of Stability from the provided temperature coefficient, load and line regulation and calibration interval information.