using external trigger
2015, 2015-P, 2016, 2016-P
WALLISP
Country: United Kingdom
Post by WALLISP » October 6th, 2011, 7:16 am
I am trying to configure the 2016 THD multimeter to take a number of voltage measurements after an external trigger is applied.
I am having difficulty working out which GPIB commands to use.
I need to use the external trigger and then wait a number of milliseconds then take upto 100 readings and then retrieve then over the gpib bus.
I am using the sample code on page 241 of the manual as a starting point.
But this code does not appear to work when i use EXT Trigger. When i send the READ? command i always get the -214 error message 'Trigger Deadlock'.
I
Re: using external trigger
Post by Dale C » October 10th, 2011, 12:09 pm
:TRIG:DELAY 1
This would set up a one second delay for each reading
:TRIG:COUNT 100
This would set up for 100 readings.
:TRIG:SOURCE EXT
This would configure the unit for an external trigger. Meaning that it now requires an external trigger to take a reading.
Sending :READ? now would result in TRIGGER DEADLOCK.
You would have to first apply the external trigger and then use :FETCH?
You would need to do this 100 times.
Post by Dale C » February 14th, 2012, 5:58 am
READ? is an
:ABORT
:INIT
:FETCH?
When the unit is set for external trigger the unit needs the external trigger to take a reading.
But the :FETCH? is asking for a reading with no trigger.
So TRIGGER DEADLOCK.
It cannot send a reading because it is waiting for the external trigger.