248 not outputting low voltage
devito
Post by devito » May 9th, 2014, 4:35 pm
I am trying to use the Keithley 248 to sweep voltage from 0-1800 V in a semi-logarithmic way. Currently, I'm using the face panel to manually set the voltage and am measuring the output. The supply tracks well above 50V, but I cannot get the supply to output voltages lower than ~44. From 50 V, I can slowly step down to 7V, but cannot go any lower than that. Any ideas as to why that is? I can't find anything in the manual about issues going low. Would it be any different if I use a DC power supply (0-10V) to set the voltage or if I program it using the GPIB?
Re: 248 not outputting low voltage
Post by Andrea C » May 9th, 2014, 5:26 pm
Are either of the filters enabled?
From the specifications/data sheet, the voltage set accuracy has a 5volt adder when the filters are in use.
Voltage Set Accuracy = 0.01% of setting + 2.5v + 5v if filter in use
Also note that the voltage regulation specs apply for values of 25 volts and higher (full load) or 50 volts and higher if no load.
devito
Post by devito » May 13th, 2014, 10:09 am
Thanks for your response Andrea!
I had not enabled the filters, but I was not aware of the regulation specs for no load vs load. I'm planning on getting a tee divider so that I can put a load on and still drive what I need. But I still want this supply to be able to give me voltages under 25V, but it looks like I'll need to create a voltage divider network in order to get below 25V.