KUSB-488A not work in Agilent IO

Post Reply
rala10
Posts: 2
Joined: May 8th, 2014, 4:34 am
Country: Germany

KUSB-488A not work in Agilent IO

Post by rala10 » May 8th, 2014, 4:37 am

Hello,

There is a problem in the communication between Agilent N8974A and PC via KUSB 488A. I install the converter's driver. But I can not detect it from Agilent IO.Even the TETEST.exe can not detect the board (No Board Found on the system).

By the way, there are two signa on the top of board. one is READY (orange), one is ACTIVE (blind). The state is OK? or where can I find the instruction. Many thanks.

Stuart M
Keithley
Keithley
Posts: 413
Joined: September 22nd, 2010, 1:42 pm
Country: United States

Re: KUSB-488A not work in Agilent IO

Post by Stuart M » May 8th, 2014, 6:54 am

I would do the following:

Check your c:\windows\system32\gpib32.dll file properties and see what the company and file version is....if Agilent, for instance, then that would explain your problem.

In all events, try the following:

1. Uninstall any Agilent GPIB driver (if present).

2. Uninstall the Keithley KUSB-488A driver

3. Search your system and remove any gpib-32.dll files (esp. from the :\windows\system32\ and c:\windows\syswow64 directories)

4. Reboot system

5. Download and install the latest KUSB-488A driver from Keithley's website which is appropriate for your hardware. It is possible you have an older driver version.
http://www.keithley.com/support/data?asset=53029

6. Login in and then connect your KUSB-488A.
If the device is not automatically recognized and connected, then:
Go to Control Panel -> Device Manager
Select "Action" from the top menu.
Then select "Scan for hardware changes"....Wait for device driver to completely install.
If successful, should appear in Device Manager listed under GPIB Interface.

7. If successful, connect an instrument and run the driver tests below to confirm presence of device and instrument:
Run the Keithley Instruments -> GPIB-488 -> GPIB Diagnostic (should display Board Id: GPIB0). If this works then use CEC -> TRTEST.EXE to talk to your instrument.

8. You should also be able to run the diagnostics that come with your card to confirm.
Finally, you should be able to run Agilient IO Connection Expert and see the GPIB device and instrument, if connected.

rala10
Posts: 2
Joined: May 8th, 2014, 4:34 am
Country: Germany

Re: KUSB-488A not work in Agilent IO

Post by rala10 » May 8th, 2014, 8:27 am

Thank you for your reply. But I am using the company computer. I do not have admin account. So I can not remove and setup software easily.

'Check your c:\windows\system32\gpib32.dll file properties and see what the company and file version is....if Agilent, for instance, then that would explain your problem.'

It is Agilent. So is it the mean the problem is in here? I really need to do the following setps?

Would you please give me your contact? I write email to your Germany office, but until now, I do not get feedback.

Stuart M
Keithley
Keithley
Posts: 413
Joined: September 22nd, 2010, 1:42 pm
Country: United States

Re: KUSB-488A not work in Agilent IO

Post by Stuart M » May 8th, 2014, 8:45 am

Yes..it looks like you do have an Agilent GPIB driver installed.
You will have to work with your company to have an admin remove the Agilent driver.
Also, uninstall the Keithley GPIB driver and reboot.
Then check the system after it is uninstalled for gpib-32.dll files.
Then install the newest Keithley GPIB driver.