Measurement of Micro Ohm with DAQ6510 - Module 7700
lucianogarc
Country: Brazil
Post by lucianogarc » September 8th, 2019, 5:30 pm
Dear Tektronix Support crew.
Hereby this contact, I really need the support to clarify any doubt that I have evolving the module 7700 - DAQ6510 using the function of 4-wire measurement.
Throughout many manuals, videos on youtube as well as datasheet I looked for this info in detail but I am still on doubt and I am avoiding to make any connection.
The structure of the test I need to do with my product correlate Temperature and Resistance. I already connect all Thermocouple (6 points) and to conclude I also need to connect the resistance (4 points).
Note: Resistors will be in 4-wire using channel 1-11(paired), 2-12( (paired) , 3-13( (paired) and 4-14( (paired) . Thermocouples will use channel 5, 6, 7, 8, 9, 10.
I share the schematic pattern that I used to connect the resistors and one photo of mine 7700 module with all connections that I made on my understanding of the pattern.
Doubt 1) Do I really need to connect wires between channel 1 and 11 or the system pairs it automatically when I select 4-wire function? You can see from the photo that I made it with red wires.
Doubt 2) Do I need to connect any other channel of the 7700 module with these set of channels that I used for resistance? If yes, can you send me any pattern or photo detailing each wire?
Doubt 3) DO you have any other pattern (or a real photo) of a 7700 module configured to work with 4-wire measurement? If no, can you develop a pattern using channels 1, 2, 3, 4 in a 4-wire function using the file attached?
Doubt 4) Using Module 7700 am I correct to think that I will connect only 2 wires between the resistor and all of the connection that characterize the 4-wire method will be internally in the 7700 module?
Any help related to my doubts will be a good advance on my tests.
Thank you!!
My Actual Configuration.pdf
(281.31 KiB) Downloaded 113 times
Module 7700.pdf
(161.65 KiB) Downloaded 109 times
4-wire assembly - Pattern followed.JPG (89.54 KiB) Viewed 930 times
Re: Measurement of Micro Ohm with DAQ6510 - Module 7700
Post by Andrea C » September 10th, 2019, 11:03 am
If I understand correctly, you are presently using jumper wires inside the 7700 to connect the HI of chan 1 and 11 together and the LO of chan 1 and 11 together. And then you run a single pair of wires outside of the 7700 to the resistor DUT.
This is not giving you lead wire error elimination benefit of 4-wire ohms.
For a four wire ohms channel on the 7700, you should run wires from both of the MUX card channels to the DUT. Bring four wires out:
4Wire Ohms
lucianogarc
Country: Brazil
Post by lucianogarc » April 16th, 2020, 5:58 pm
Thank you so much for your reply!
I will have to rework my MUX7700 instalation to check any difference.
Using 4 Wires, Channel 1 - 11.
Should I use 1 shielded cable with 4 wires inside to each DUT resitor (instead of jumping, I will have 4 wires coming out of the MUX ??
or it must be 2 shielded cables with 2 wires inside for each DUT resistor?
Let make it clear to me.
Post by Andrea C » April 23rd, 2020, 5:11 am
Given the physical locations of the screw terminals for the 4W channel pairs on the 7700 and the cable routing in the card, a cable with 4 conductors inside will be very difficult to use. Go with two cables, each with two conductors.