Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Low Resistane Measuring

Post Reply
trm60101
Posts: 1
Joined: July 9th, 2012, 7:16 pm
Country: United States

Low Resistane Measuring

Post by trm60101 » July 9th, 2012, 8:11 pm

My client currently uses a Keithley Model 7002 Mainframe, a Model 487 - Ammeter / Voltage Source and Model 7011 Multiplexer / Relay Cards (modified for form C switching) for high resistance (SIR) resistance testing. (see attached diagram). They need to do low resistance testing typically
2 to 10 ohms with a 20 volt bias voltage and a 12 volt test voltage.
Can they use the current 7002 Mainframe and 7011 Card set up and replace the 480 - Ammeter / Voltage Source with another instrument(s)
to accomplish this? If so , how would it be connected?
Attachments
Keithly SIR Set Up01.pdf
SIR Set UP
(229.82 KiB) Downloaded 1331 times

Mark W
Keithley Applications
Keithley Applications
Posts: 14
Joined: July 9th, 2012, 5:52 am
Country: United States

Re: Low Resistane Measuring

Post by Mark W » July 11th, 2012, 7:56 am

They can use the 7002 and 7011 as well as the 480. There should be no need to replace the -. I have attached a diagram to illustrate the connections.

Mark - Applications