Hi,
Is it possible to program the 2182A to perform a number of DC measurements ( let say 20) and then recover the measurements from the buffer. I would like to configure the machine for doing the measurements and later recover them instead of doing individual measurement/reading.
In the manual is written:
:TRAC:POIN 20 Set buffer size to 20.
:TRAC:FEED SENS Set to store raw readings.
:TRAC:FEED:CONT NEXT Start storing readings.
:TRAC:DATA? Request all stored readings.
Does it mean that the 2182A is going to perform 20 measurements? Or does it simply save the next 20 measurements in memory?
Thanks
Is it possible to program the 2182A to perform a number of DC measurements ( let say 20) and then recover the measurements from the buffer. I would like to configure the machine for doing the measurements and later recover them instead of doing individual measurement/reading.
In the manual is written:
:TRAC:POIN 20 Set buffer size to 20.
:TRAC:FEED SENS Set to store raw readings.
:TRAC:FEED:CONT NEXT Start storing readings.
:TRAC:DATA? Request all stored readings.
Does it mean that the 2182A is going to perform 20 measurements? Or does it simply save the next 20 measurements in memory?
Thanks
-
- Keithley Applications
- Posts: 2841
- Joined: June 10th, 2010, 6:22 am
- Country: United States
- Contact:
Re: keithley 2182A: Continues dc measurement
One more command is required.
Place :INIT between
:TRAC:FEED:CONT NEXT
and
:TRACE:DATA?
The :INIT will start the data taking process.
Then the Model 2182A will take 20 readings and store them in the buffer when the :INIT command is received.
Then :TRACE:DATA?
will retrieve the 20 readings when the 2182A is addressed to talk.
Place :INIT between
:TRAC:FEED:CONT NEXT
and
:TRACE:DATA?
The :INIT will start the data taking process.
Then the Model 2182A will take 20 readings and store them in the buffer when the :INIT command is received.
Then :TRACE:DATA?
will retrieve the 20 readings when the 2182A is addressed to talk.
Re: keithley 2182A: Continues dc measurement
Shall I introduce a delay between :INIT and :TRACE:DATA?. to let some time to the 2182A to take the measurements?
Thanks
Thanks
-
- Keithley Applications
- Posts: 2841
- Joined: June 10th, 2010, 6:22 am
- Country: United States
- Contact:
Re: keithley 2182A: Continues dc measurement
Yes. That would be a good idea.
The amount of delay depends on how the 2182A reading rate is congfigured.
The amount of delay depends on how the 2182A reading rate is congfigured.
Re: keithley 2182A: Continues dc measurement
Hello,
I have followed the indications and I have add the command “INIT”. However I do not get any response from the device.
I have moved to the VIs provided by Keithley but I could not make it work. I did proceed as follow:
Frame1:
Measurement mode VI (Selected chanel 1)
DCV config VI (default values except NLPC=0.1)
Buffer Config VI------- Here I have modified the VI. In the original VI is written “:Trac:Feed:Cont Next;” and I change it by “:Trac:Feed Sens;:Trac:Feed:Cont Next;” (Options: Configure Buffer, Next and 5 Points)
Initiate/Abort VI (Option: Init One cycle)
Frame2:
Delay of 20 seconds To be sure that the machine can take the readings.
Frame3:
Buffer Read VI
Right now, I don’t know what I am doing wrong. Could you give me any indication.
Thanks
I have followed the indications and I have add the command “INIT”. However I do not get any response from the device.
I have moved to the VIs provided by Keithley but I could not make it work. I did proceed as follow:
Frame1:
Measurement mode VI (Selected chanel 1)
DCV config VI (default values except NLPC=0.1)
Buffer Config VI------- Here I have modified the VI. In the original VI is written “:Trac:Feed:Cont Next;” and I change it by “:Trac:Feed Sens;:Trac:Feed:Cont Next;” (Options: Configure Buffer, Next and 5 Points)
Initiate/Abort VI (Option: Init One cycle)
Frame2:
Delay of 20 seconds To be sure that the machine can take the readings.
Frame3:
Buffer Read VI
Right now, I don’t know what I am doing wrong. Could you give me any indication.
Thanks
-
- Keithley Applications
- Posts: 2841
- Joined: June 10th, 2010, 6:22 am
- Country: United States
- Contact:
Re: keithley 2182A: Continues dc measurement
Try these SCPI command codes.
*RST
:TRACE:CLEAR
:SAMPLE:COUNT 20
:TRACE:POINTS 20
:TRACE:FEED SENS
:TRACE:FEED:CONT NEXT
:INIT
Wait about 6 or 7 seconds for it complete.
:TRACE:DATA?
Now read the data.
*RST
:TRACE:CLEAR
:SAMPLE:COUNT 20
:TRACE:POINTS 20
:TRACE:FEED SENS
:TRACE:FEED:CONT NEXT
:INIT
Wait about 6 or 7 seconds for it complete.
:TRACE:DATA?
Now read the data.
Re: keithley 2182A: Continues dc measurement
Thank you it works perfect.
It was missing :SAMPLE:COUNT 20 command
It was missing :SAMPLE:COUNT 20 command
-
- Posts: 2
- Joined: October 7th, 2014, 4:34 am
- Country: India
Re: keithley 2182A: Continues dc measurement
Hi,
I am need to load 1000 data with the trigger and read the data stored in Trace by initiating it. I followed the steps given here (in the previous posts). But I could not read more than 60 data from the buffer. The scenario is, when I set the Sample count >60 (say 100) and the Trace Store points as 100, I end up with Timeout error (even with delay set to 20s) and could not read. Instead if I try to set the Trace Store Points<=60, I am able to read the 60 values form the buffer.
Can anyone please help me and let me know where I am going wrong and get the data read from the buffer?
Thanks in Advance.
Regards,
Shenbaga Priya
I am need to load 1000 data with the trigger and read the data stored in Trace by initiating it. I followed the steps given here (in the previous posts). But I could not read more than 60 data from the buffer. The scenario is, when I set the Sample count >60 (say 100) and the Trace Store points as 100, I end up with Timeout error (even with delay set to 20s) and could not read. Instead if I try to set the Trace Store Points<=60, I am able to read the 60 values form the buffer.
Can anyone please help me and let me know where I am going wrong and get the data read from the buffer?
Thanks in Advance.
Regards,
Shenbaga Priya
-
- Keithley Applications
- Posts: 2841
- Joined: June 10th, 2010, 6:22 am
- Country: United States
- Contact:
Re: keithley 2182A: Continues dc measurement
Try this code.
*RST
:TRACE:CLEAR
:VOLT:NPLC 0.01
:SAMPLE:COUNT 1000
:TRACE:POINTS 1000
:TRACE:FEED SENS
:TRACE:FEED:CONT NEXT
:INIT
Wait about 20 seconds for it complete.
:TRACE:DATA?
Now read the data.
This worked fine in my Labview 2012 program.
Maybe in your program the timeout setting is not set long enough?
*RST
:TRACE:CLEAR
:VOLT:NPLC 0.01
:SAMPLE:COUNT 1000
:TRACE:POINTS 1000
:TRACE:FEED SENS
:TRACE:FEED:CONT NEXT
:INIT
Wait about 20 seconds for it complete.
:TRACE:DATA?
Now read the data.
This worked fine in my Labview 2012 program.
Maybe in your program the timeout setting is not set long enough?
-
- Posts: 2
- Joined: October 7th, 2014, 4:34 am
- Country: India
Re: keithley 2182A: Continues dc measurement
Hi Dale C,
Thanks for the response. I tried both in LabVIEW 2010 and 2012 with timeout configured for more than 30s. The measurement delay is also configured 25s.
I am ending in timeout error. How can I ensure that the the trigger operation is complete. Can using OPC query be helpful in this case? Or I could try with
:INIT
Wait for 30s
:ABORt
:TRAC:DATA?
Thanks in advance.
Regards,
Shenbaga Priya
Thanks for the response. I tried both in LabVIEW 2010 and 2012 with timeout configured for more than 30s. The measurement delay is also configured 25s.
I am ending in timeout error. How can I ensure that the the trigger operation is complete. Can using OPC query be helpful in this case? Or I could try with
:INIT
Wait for 30s
:ABORt
:TRAC:DATA?
Thanks in advance.
Regards,
Shenbaga Priya
-
- Keithley Applications
- Posts: 2841
- Joined: June 10th, 2010, 6:22 am
- Country: United States
- Contact:
Re: keithley 2182A: Continues dc measurement
You could do *OPC to do this.
You could put in a hard delay, like you show below.
You could have the unit SRQ on buffer full.
Either method would do it.
Most users like the SRQ on buffer full.
Just send :STAT:MEAS:ENAB 512 for Buffer full
and *SRE 1
Then have your program wait for an SRQ.
That should do it.
You could put in a hard delay, like you show below.
You could have the unit SRQ on buffer full.
Either method would do it.
Most users like the SRQ on buffer full.
Just send :STAT:MEAS:ENAB 512 for Buffer full
and *SRE 1
Then have your program wait for an SRQ.
That should do it.