I am attempting to command the Keithly-237 via GPIB using C# and .NET. I wish to
1) read its identification string
2) command it to Source I Measure V as DC
3) set it to source (and hold) one specific current (say 23.45 microAmps)
I do not need to sweep through a range or measure anything.

I understand the version of GPIB is old and will not recognize queries like *IDN?, but I cannot find specifically from the manual how to do the above.

Any help is appreciated. Thank you.

Alan

Dale C
Keithley Applications
Keithley Applications
Posts: 2847
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: Keithley-237 GPIB commands

Post by Dale C » August 20th, 2013, 8:02 am

1. To read the ID stirng send U0C then read it.
2. To source I and measure V in DC mode send F1,0X.
3. To source 23.45uA send B23.45e-6,6,0X
Here is the link to the manual that shows all he DDCs(Device Dependent Commands)