connect 6221 to several 2182 at the same time

2182 Series Support
Post Reply
czhangfdu
Posts: 1
Joined: October 16th, 2013, 9:08 am
Country: China

connect 6221 to several 2182 at the same time

Post by czhangfdu » October 16th, 2013, 9:13 am

Hello,
I was wandering if 6221 can be connected to several 2182 at the same time for delta mode?
Thanks!
Cheng

Kenneth P
Keithley Applications
Keithley Applications
Posts: 212
Joined: October 25th, 2010, 1:31 pm
Country: United States
Contact:

Re: connect 6221 to several 2182 at the same time

Post by Kenneth P » October 21st, 2013, 8:49 am

Hello Cheng,

It is not possible to connect more than one 2182A to the 6221 while using Delta mode.The Delta mode operation uses the 6221 as the master and readings are taken and displayed using the 6221 and it is designed to only communicate with only one 2182A