Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Possible Current Compliance on 4200-SCS w/ 4224-RPM ?

Post Reply
rododendron
Posts: 3
Joined: November 25th, 2013, 6:48 am
Country: Germany

Possible Current Compliance on 4200-SCS w/ 4224-RPM ?

Post by rododendron » November 25th, 2013, 6:55 am

I would like to know if is there a possibility to use current compliance with setup 4200-SCS + 2x 4224-RPM (e.g. for pulse measurements) and if so, how can be this set. We haven't figured out. Thank you!

Andrea C
Keithley Applications
Keithley Applications
Posts: 1512
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: Possible Current Compliance on 4200-SCS w/ 4224-RPM ?

Post by Andrea C » November 27th, 2013, 5:22 am

Yes, it is possible for the RPM to apply a current limiting compliance.

To access this feature, you will need to make use of a UTM that is calling the Set_RPM_ICompliance function.

First step is to characterize the compliance limiting of each RPM. A SMU is connected to the output of the RPM for this step. There is a User Library and module for performance of this step.
User Library: RPM_ILimit_Control
User Module: Do_RPM_ILimit_Cal

Read the test description carefully for how to connect the SMU during this step of the process.

Once the characterization is done, then other UTMs can be used to apply voltage pulses with a programmed current limit. For an example of this, see the following:
User Library: PMU_PCRAM_ulib
User Module: pram_pulse_ilimit

Andrea

rododendron
Posts: 3
Joined: November 25th, 2013, 6:48 am
Country: Germany

Re: Possible Current Compliance on 4200-SCS w/ 4224-RPM ?

Post by rododendron » December 2nd, 2013, 4:42 am

Dear Andrea,

thank you for your reply. Unfortunately we don't have any SMU module, only one PMU + 2x RPM. If I understand correctly, it's not possible to use current compliance without an SMU.

thanks.

zvikaef
Posts: 16
Joined: June 9th, 2016, 2:57 am
Country: Israel

Re: Possible Current Compliance on 4200-SCS w/ 4224-RPM ?

Post by zvikaef » October 16th, 2016, 2:40 am

Hi Andrea.
when using above described procedure, what is the min. value of compliance that can be programed?

Andrea C
Keithley Applications
Keithley Applications
Posts: 1512
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: Possible Current Compliance on 4200-SCS w/ 4224-RPM ?

Post by Andrea C » October 17th, 2016, 6:07 am

The 4225-RPM current compliance settings are from about 10uA to 10mA.
RPM_limit_Control.PNG
RPM current limit calibration
RPM_limit_Control.PNG (114.1 KiB) Viewed 19170 times

Post Reply

Return to “4200-SCS Semiconductor Parameter Analyzer”

Who is online

Users browsing this forum: No registered users and 0 guests