How fast can you query the current measured by Model 2230-30
2200-20-5, 2200-30-5, 2200-32-3, 2200-60-2, 2200-72-1, 2220-30-1, 2230-30-1
Kevin62
Country: Canada
Post by Kevin62 » February 6th, 2014, 4:06 pm
How fast can you query the Model 2230-30-1 power supply using the command "MEASure:CURRent? CH1 "
We are finding that there needs to be a delay between the VISA Write and the VISA Read in order for the power meter to return a current value consistently.
If the delay is too short, i.e. 50 ms the power meter returns "zero" on every 2nd execution in a command loop.
If the delay is 200 ms the power meter returns the command consistently in the command loop but that meant it was almost a second for every query which is slow.
What is the expected amount of time (delay) for the power meter to process the command "MEASure:CURRent? CH1" consistently ?
Regards,
Kevin
Re: How fast can you query the current measured by Model 223
Post by Dale C » February 7th, 2014, 12:11 pm
Hello Kevin,
The 2200 series supplies are basically slow.
I would expect about 150msec for responses.
Return to “Series 2200 Programmable DC Power Supplies”