Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

question on 6221 Delta Ohms reading

6220, 6221 Current Sources
Post Reply
hachiko
Posts: 18
Joined: May 15th, 2013, 6:02 am
Country: China

question on 6221 Delta Ohms reading

Post by hachiko » October 9th, 2014, 5:54 am

I just found a problem when I measured Hall resistance. With Delta Ohms reading, 6221(together with 2182A) always gives positive values (Ohms). As you know, Hall resistance should be negative on reversing magnetic field.
Does 6221 automatically make the absolute values when Delta measurement unit is Ohm? And Is there the same problem with differential conductance (dV/dI)?

Dale C
Keithley Applications
Keithley Applications
Posts: 2849
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: question on 6221 Delta Ohms reading

Post by Dale C » October 9th, 2014, 6:09 am

Delta mode will always show a positive reading.
This is because the delta mode uses two voltage readings to calculate the delta mode reading.
The delta mode sources a positive current and takes a voltage reading.
The it sources a negative current of the same magnitude and takes another voltage reading.
Then some algorithm that is described in the 6221 and 2182A manuals is performed.
Then the delta mode reading is displayed.

hachiko
Posts: 18
Joined: May 15th, 2013, 6:02 am
Country: China

Re: question on 6221 Delta Ohms reading

Post by hachiko » October 9th, 2014, 6:31 am

Do you mean even if the Delta measurement unit is Volt, it gives always positive reading?
If so, that will be a pity. Because, if one has the connection order of I+ V- V+ I-, one would not get the correct signal of this order.
Last edited by hachiko on October 9th, 2014, 6:52 am, edited 2 times in total.

hachiko
Posts: 18
Joined: May 15th, 2013, 6:02 am
Country: China

Re: question on 6221 Delta Ohms reading

Post by hachiko » October 9th, 2014, 6:47 am

See the attached figure from the manual.
The 2182A A reading could be negative at I-high, and B reading could be positive at I-Low, if you have the 4 contacts on a resistance in the order of I- V+ V- I+. So with this Delta algorithm, the Delta voltage reading could be negative.
Delta.PNG
1
Delta.PNG (35.43 KiB) Viewed 19297 times
As for Delta Ohm reading, according to the manual (see the snapshot from the manual), it could be negative too. I don't know whether there is any other details in the manual.
Ohm reading.PNG
Ohm reading.PNG (18.25 KiB) Viewed 19297 times

Dale C
Keithley Applications
Keithley Applications
Posts: 2849
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: question on 6221 Delta Ohms reading

Post by Dale C » October 9th, 2014, 12:59 pm

:oops:
I guess I have never considered using negative values for the high value.
So you are correct.
I have never run into such an application. Therefore, I have never seen a negative value for delta.
Thanks for pointing it out.
Not sure how that relates to your HALL effect issue.

Post Reply

Return to “Current Sources”

Who is online

Users browsing this forum: No registered users and 0 guests