Stop at compliance without sweep

Post by noddle » April 1st, 2011, 12:03 am

Hi,
I use Keithley 2410 source-meter to do TDDB measurement. I want this source-meter stop at current compliance.

I have tried to put this code

:STAT:MEAS:ENAB 16384
*SRE 1
:ABOR
:STAT:MEAS?
*SRE 0

but it didnot work. Any help would be greatly appreciated.

Dale C
Keithley Applications
Keithley Applications
Posts: 2849
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: Stop at compliance without sweep

Post by Dale C » April 1st, 2011, 7:34 am

Are you doing a sweep?
If so then use the compliance abort command.
:SOURCE:SWEEP:CABORT (NEVER,EARLY or LATE).
If not using a sweep then you could also use the trip command.
:SENSE:CURRENT:PROTECTION:TRIPPED?
This would return a yes(1) or no(0) when addressed to talk.