Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Measuring nanovolts DC with Keithley 2002

2182 Series Support
Post Reply
Heusler
Posts: 6
Joined: September 6th, 2015, 10:32 am
Country: Russian Federation

Measuring nanovolts DC with Keithley 2002

Post by Heusler » October 23rd, 2015, 8:57 am

Hello,

Is it possible to reliably measure nanovolts DC with Keithley 2002? According to its manual it has resoultion down to 1 nV.
However, on the lowest scale (200 mVDC) and integration time 10 PLC, slowly changing DC voltages about 10 µV are measured quite noisy, signal-to-noise ratio is about 5%.
When the same line used for measuring with a different nanovoltmeter the measurements, are far less noisy, about 1%.

brian.d.smith
Keithley Applications
Keithley Applications
Posts: 402
Joined: June 10th, 2013, 6:00 am
Country: United States

Re: Measuring nanovolts DC with Keithley 2002

Post by brian.d.smith » October 23rd, 2015, 10:11 am

You must be careful with microvolts that are thermally induced. You should use solid copper everywhere in your adapters and cables.

Shielding is required from the front panel of the 2002 all the way to your device.

Nanovoltmeters have these features built in.

Heusler
Posts: 6
Joined: September 6th, 2015, 10:32 am
Country: Russian Federation

Re: Measuring nanovolts DC with Keithley 2002

Post by Heusler » October 23rd, 2015, 4:29 pm

thank you for the answer!
But what is the real limit of this model?
What is the minimal signal-to-noise ratio at 10 nV, 100 nV, 1 mV DC signals?

brian.d.smith
Keithley Applications
Keithley Applications
Posts: 402
Joined: June 10th, 2013, 6:00 am
Country: United States

Re: Measuring nanovolts DC with Keithley 2002

Post by brian.d.smith » October 24th, 2015, 7:49 am

Please see page 2 of the attached customer specification. RMS noise is specified as ppm.
SPEC-2002 (I - Nov 2009)(2).pdf
(237.6 KiB) Downloaded 1233 times

Post Reply

Return to “Nanovoltmeters”

Who is online

Users browsing this forum: No registered users and 0 guests