I need to use the "Trip Relay mode" function but I found that in my 2303 it doesn't work because I have the firmware version A04.
Where can I find the firmware A06 mentioned in the manual that enable this function?
In the section dedicated to the model 2303 of your site there is only the Flashwizard program but no firmware.
Dale C
Keithley Applications
Country: United States
Contact:
Re: 2303 Trip Relay mode
Post by Dale C » November 18th, 2015, 7:40 am
I do not see an A06 revision.
The revision A10 is attached. It is the latest.
2303-803A10.x
(657.36 KiB) Downloaded 1175 times
picburner
Post by picburner » November 18th, 2015, 10:21 am
Many thanks for your prompt response.
I make the upgrade now and try the new features.
Regards.