Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

2303 Trip Relay mode

2302, 2303, 2303-PJ, 2304, 2306, 2306-PJ, 2306-VS, 2308
Post Reply
picburner
Posts: 14
Joined: September 10th, 2014, 10:15 am
Country: United Kingdom

2303 Trip Relay mode

Post by picburner » November 16th, 2015, 6:05 am

I need to use the "Trip Relay mode" function but I found that in my 2303 it doesn't work because I have the firmware version A04.
Where can I find the firmware A06 mentioned in the manual that enable this function?
In the section dedicated to the model 2303 of your site there is only the Flashwizard program but no firmware.

Dale C
Keithley Applications
Keithley Applications
Posts: 2847
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: 2303 Trip Relay mode

Post by Dale C » November 18th, 2015, 7:40 am

I do not see an A06 revision.
The revision A10 is attached. It is the latest.
2303-803A10.x
(657.36 KiB) Downloaded 1177 times

picburner
Posts: 14
Joined: September 10th, 2014, 10:15 am
Country: United Kingdom

Re: 2303 Trip Relay mode

Post by picburner » November 18th, 2015, 10:21 am

Many thanks for your prompt response.
I make the upgrade now and try the new features.
Regards.

Post Reply

Return to “Series 2300”

Who is online

Users browsing this forum: No registered users and 1 guest