Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Two DMM7510s making simultaneous measurements?

DMM7510 7½-Digit Graphical Sampling Multimeter
Forum rules
<t></t>
Post Reply
aVictor
Posts: 1
Joined: January 29th, 2016, 4:47 am
Country: United States

Two DMM7510s making simultaneous measurements?

Post by aVictor » January 29th, 2016, 4:58 am

We have two DMM7510 units. We need to get the two units making simultaneous VDC measurements of two different circuits.

I tried Kickstart, but it seems to be limited to a single DMM7510.

Is there a way to synchronize two DMM7510 and get continuous VDC samples from both units?

There's no voltage trigger in this setup, but I would settle for stopping VDC measurements on both units simultaneously and writing down the VDC measurement value. If that's the route, how can I stop both units at the same time?

Joseph V
Keithley Applications
Keithley Applications
Posts: 107
Joined: July 1st, 2014, 10:31 am
Country: United States

Re: Two DMM7510s making simultaneous measurements?

Post by Joseph V » February 19th, 2016, 11:34 am

If you want to synchronize two DMM7510's together the simplest way is to trigger the instruments using TSP-Link.
Attached are two .tsp files which set up each instrument for synchronizing measurements. Note this is a very basic application which has some variation between reading times however for the most part each instrument is acquiring a reading ~6 microseconds apart from one another.
Attachments
UnitSyncNode2.tsp
(1.27 KiB) Downloaded 1100 times
UnitSyncNode1.tsp
(1.29 KiB) Downloaded 1103 times

Post Reply

Return to “DMM7510 Graphical Sampling DMM”

Who is online

Users browsing this forum: No registered users and 0 guests