Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Synchronized Time Reference

Post Reply
Matthew V
Posts: 1
Joined: October 16th, 2018, 12:08 pm
Country: United States

Synchronized Time Reference

Post by Matthew V » October 16th, 2018, 12:54 pm

I am performing an experiment where two currents are measured over time using Keithley Model 6514 electrometers. I don’t necessarily need to measure both currents at the precisely the same time; however, I do need the timestamp associated with each measurement to be in reference to the same time. I am operating at 6 PLC for this experiment, and I am using LabVIEW. Are there any software or hardware solutions? Thanks.

Andrea C
Keithley Applications
Keithley Applications
Posts: 1244
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: Synchronized Time Reference

Post by Andrea C » December 4th, 2018, 5:13 pm

Are you triggering N measurements on each unit? Or doing a software loop and one reading per iteration?

If triggering, it would be advised to put a trigger link cable between the two units and adjust the trigger of N readings so that one meter outputs a "measure now" signal and the other one waits for this.

Post Reply

Return to “Low Current Measurements”

Who is online

Users browsing this forum: No registered users and 1 guest