Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.
Trigger Model - Sine Wave of Current
Models 2450, 2460, 2461, 2470
Post Reply
Andrea C
Keithley Applications
Country: United States
Contact:
Post by Andrea C » June 10th, 2021, 7:50 am
Using 2450 as a current source with voltage limiting, we can generate slow sine waves.
Here is a scope shot of a 36 point per cycle, 10Hz, 10mA peak-peak sine into 500Ω:
scope shot - 10mA pp into 500Ω
On the instrument:
Reading Table View on the Instrument
Attached are some TSP files for the capability.
KI2450_current_sine_wave_functions_define.tsp
Defines helper functions. Load them to runtime memory.
(3.72 KiB) Downloaded 3 times
KI2450_current_sine_wave_functions_use.tsp
Use our functions - VISA write these from your favorite programming lang
(1.53 KiB) Downloaded 2 times