Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

7708 module 4-wire TEMP measurements (PT100)

7001, 7002 and Switch Cards
Post Reply
Dale C
Keithley Applications
Keithley Applications
Posts: 2847
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: 7708 module 4-wire TEMP measurements (PT100)

Post by Dale C » June 13th, 2011, 7:39 am

Yes. Each channel on the 7708 card can have it's own function.
For example.
Channel 1 could be Temperature with type T thermocouple.
Channel 2 could be a four wire RTD (Please note this would gang channel 22 for the four wire connection).
This is simple to set up in the Excelinx-1A software
Look at the DMM CONFIG sheet. Line 36.
Set channel 1 for function = Temp
Set RANGE to T
Set Option to INT
Now set channel 2 to Temp.
Set Range to PT100
That should do it.