Dear Sir or Madam,
we are using a DMM 2701/E together with a switching module 7700.
In the Model 2701 Users Manual - Chapter 3-10 (Basic DMM Operation) there is some information about the connection of the Model 7700 switching module:
Maximum input voltage: 300VDC or RMS, 1A switched, 60W, 125VA
Why is the maximum dc voltage level the same as RMS voltage level?
300VRMS -> 424,26Vpeak
Can we connect a DC voltage of 400VDC if we are switching only 40uA current?
Please give me a short explanation.
Thanks,
W. Schafroth / SET GmbH
Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.
7700 Switching Module - Max. DC voltage
-
- Keithley Applications
- Posts: 2849
- Joined: June 10th, 2010, 6:22 am
- Country: United States
- Contact:
Re: 7700 Switching Module - Max. DC voltage
The maximum voltage on the 7700 card is 300V. Do not apply more than 300V or damage could occur.
even if switching with a low current. This is not a good idea.
even if switching with a low current. This is not a good idea.