Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

2306 Sense Impedance?

2302, 2303, 2303-PJ, 2304, 2306, 2306-PJ, 2306-VS, 2308
Post Reply
IvorBowden
Posts: 6
Joined: May 1st, 2013, 6:44 pm
Country: United States

2306 Sense Impedance?

Post by IvorBowden » May 1st, 2013, 6:47 pm

What is the impedance of the 2306 sense inputs? Specifically, how much series resistance can I have in the sense lines without affecting accuracy?

Dale C
Keithley Applications
Keithley Applications
Posts: 2847
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: 2306 Sense Impedance?

Post by Dale C » May 2nd, 2013, 8:16 am

The 2306 has an effective 10Meg input impedance on the sense lines. The 2306 has a .05% accuracy spec.

With 500ohm in each sense lead, the accuracy would be degraded by 1K/10Meg or .0001 or .01%.

But

In reality, the +sense lead has a 10Meg to internal ground, the –sense is high impedance.
So only resistance in the +sense lead will affect accuracy. So 1K in the +sense lead will affect accuracy by 1K/10Meg or .01%.
1K in the –sense lead will not affect accuracy. I wouldn’t go crazy on impedance on the –sense input, there is some leakage currents in the 1na range.

IvorBowden
Posts: 6
Joined: May 1st, 2013, 6:44 pm
Country: United States

Re: 2306 Sense Impedance?

Post by IvorBowden » May 2nd, 2013, 9:23 am

Thanks Dale!
Useful information for my application.
Regards,
Ivor

Post Reply