Save on Trigger example in Python
Chris B
Post by Chris B » December 21st, 2010, 11:39 am
As a recent learning project, I have written a Save on Trigger program that is compatible with Tektronix TDS2000B/C, TDS3000B/C, DPO/MSO2000/3000/4000 series scopes. Data is saved as Image files, or .CSV waveform files (viewable in Wavestar) directly to the PC running the Save on trigger application. This has been written with Python 2.7, and uses PyVISA 1.3. Due to complications with PyVISA not properly releasing my COM ports when a program completes (or crashes), I have decided to only code support for Ethernet, USB or GPIB connections.
Currently there are two parts of this program. SaveonTrig.py is the launcher, and used to loop the main program, as well as trap errors that would cause the program crash (typically issues with the VISAs resource list). SoTmain.py contains all the code that handles setting up the scope, capturing the selected data types and saving a log of the session.
SaveonTrig.py
Launch program for main SaveonTrig script
This module is used to loop the SoTmain application
when desired and provide more information
about how to use the program.  Built using
python 2.7 (http://www.python.org/)
pyvisa 1.3 (http://pypi.python.org/pypi/PyVISA/1.3)
import SoTmain
print "Welcome to the save on Trigger Data Logging Application"
print "This program is compatible with TDS2000B/C, TDS3000B/C"
print "and DPO2000/3000/4000 series Oscilloscopes with a TCPIP"
print "or USB connection. NI-VISA or TekVISA must be installed. \n"
while True:
    try:
        SoTmain.main()
    except:
        print "An unexpected error occured"
    print "\n\nWould you like to start another capture" \
          + " session? (Y/N)"
    Doagain = str(raw_input(""))
    if ('Y' not in Doagain) and ('y' not in Doagain):
        print "Good Bye!"
        break
a = raw_input("Press ENTER to exit")
SoTmain.py
Save on Trigger intended for Bench scopes.
Tested with DPO3054, MSO2024, TDS2024B, TDS3054B,
TDS2024C, DPO4032
Used USB and TCPIP Connections for the DPO scopes
Requires either TekVISA or NI-VISA installation
To change the file path edit the saveontrig.dat
file found in the directly where USBsaveontrig resides.
import sys
import struct
import os
from visa import *
from pyvisa import vpp43
from time import *
from struct import unpack
def main():
    #Initiate capture log
    capturelog = []
    # get path for files to be saved in, base file name and file format from
    # config.dat file.  Create a generic config.dat file if non exists
    if os.path.exists('saveontrig.dat') == False:
        f = open ('saveontrig.dat', 'w')
        f.write ('C:\Images\\')
        f.close()
    f = open ('saveontrig.dat', 'r')
    filepath = f.read() +  str(time()).split('.')[0] + '\\'
    f.close()
    #check if the directory for saved files is created, if not make it
    if not os.path.exists(filepath):
        os.makedirs(filepath)
        print 'Creating Directory ' + filepath
    print "Capture data and logs will be saved to " + filepath
    print "To change the root directory for captured files"
    print "edit the saveontrig.dat file located in the folder where"
    print "SaveonTrig program resides."
    # Create list of VISA objects and allow user to select which to connect to
    resource_names = []
    while True:
        try:
            find_list, return_counter, instrument_description = \
                  vpp43.find_resources(resource_manager.session, "?*")
            resource_names.append(instrument_description)
            break
        except VisaIOError:
            print "No Instruments Detected"
            raise SystemExit()
        except:
            print "Unable to communicate with VISA Driver"
            raise SystemExit()
    for i in xrange(return_counter - 1):
        resource_names.append(vpp43.find_next(find_list))
    print "\nVisa resources found:"
    for a in range(len(resource_names)):
        print a, resource_names[a]
    print "Attempting to Identify Instruments. Please Wait..."
    resource_ids = []
        interface_type, _ = \
        vpp43.parse_resource(resource_manager.session, resource_names[a])
        if interface_type == VI_INTF_ASRL:
            resource_ids.append("RS232 not Supported")
        else:
            while True:
                try:
                    tempscope = instrument(resource_names[a], timeout = 10)
                    scopename = tempscope.ask("*IDN?")
                    scopename = scopename.split(',')
                    resource_ids.append(scopename[0] + '-' + scopename[1] + \
                                        ' ' + scopename[2])
                    break
                except VisaIOError:
                    resource_ids.append(" No Instrument Found ")
                    break
                except:
                    resource_ids.appent(" Error Communicating with this device")
            del tempscope
    print "\nIdentified Instruments:"
        print a, resource_ids[a]
    # Query from user which scope to connect to
    while True:
        try:
            user_device = int(raw_input("Connect to which scope? \n"))
            if user_device in range(len(resource_names)):
                break
            else:
                print "Invalid Input, Please Try Again"
        except ValueError:
    # Connect to scope, and verify that it can be talked too
    while True:
        try:
            scope = instrument(resource_names[user_device], timeout = 15)
            scopename = scope.ask("*IDN?")
            print scopename
            break
        except:
            print "Unable to communicate over this port"
            raise SystemExit()
    # query from the user what type of capture to make
    while True:
        try:
            print "\n# Mode"
            print "0 HardCopy - Saves PNG, BMP or TIF images depending on scope"
            print "1 CSV - Savess waveform as a CSV file, viewable in Wavestar."
            print "2 Save both CSV and HardCopy"
            logmode = int(raw_input("Log which information? \n"))
            if logmode in range(3):
                break
            else:
        except ValueError:
    # query from the user how many minutes to capture data for
    while True:
        try:
            duration_time = float(raw_input("How many minutes should I run? \n"))
            break
        except ValueError:
    # Query from user if a trigger should be forced to execute periodically
    print "If no trigger is reieved, would you like to"
    Forcetrig = str(raw_input(" periodically force a trigger? (Y/N) \n"))
    if ('Y' in Forcetrig) or ('y' in Forcetrig):
        Forcetrig = True
        while True:
            try:
                Forcetrig_time = float(raw_input("How many seconds should I " \
                                        +"wait before forcing a trigger? \n"))
                break
            except ValueError:
    else:
        Forcetrig = False
    # Save current scope configuration
    print "Saving current scope setup."
    Scopesetup = scope.ask("*LRN?")
    # Determine the type of scope, and configure hard copy accordingly
    # And settings for aquiring the curve data
    if 'TDS' in scopename[1] and  '20'  in scopename[1]:
        fileformat = 'TIF'
        scope.write("header off")
        scope.write("HARDCOPY:BUTTON SAVESIMAGE")
        scope.write("HARDCOPY:FORMAT TIFF")
        scope.write("HARDCOPY:PORT USB")
        scope.write("wfmoutpre:bit_nr 8")
        Hscale = float(scope.ask("HOR:SCA?"))
        HDelay = float(scope.ask("HORizontal:DELay:POSITION?"))
        HPos = float((Hscale * 5 + float(scope.ask("HORIZONTAL:POSITION?")))/ \
                     (Hscale * 10) * 100)
    elif 'DPO' in scopename[1] or 'MSO' in scopename[1]:
        fileformat = 'PNG'
        scope.write(":header off")
        HDelay = float(scope.ask("HORizontal:DELay:TIMe?"))
        HPos = float(scope.ask("HORIZONTAL:POSITION?"))
        if '20' in scopename[1]:
            if str(scope.ask("DATA:COMP?")) != "COMPOSITE_YT":
                scope.write("DATA:COMP COMPOSITE_YT")
                print "Only Composite_YT supported for FilterVu"
    elif 'TDS' in scopename[1] and '30' in scopename[1]:
        fileformat = 'BMP'
        scope.write("HARDCOPY:PORT GPIB")
        scope.write("HARDCOPY:FORMAT BMPColor")
        HDelay = float(scope.ask("HORizontal:DELay:TIME?"))
        HPos = float((Hscale * 5 + float(scope.ask("HORIZONTAL:DELay:TIME?")))/ \
    #Generic all model configure the scope hardcopy settings
    scope.write("VERBOSE OFF")
    scope.write("HARDCOPY:LAYOUT PORTRAIT;PREVIEW 0;INKSAVER 0")
    scope.write("HARDCOPY:PALETTE NORMAL")
    scope.write("SAVE:IMAG:FILEF " + fileformat)
    recLen = scope.ask("horizontal:recordlength?")
    scope.write("data:start 1;stop " + recLen + \
                ";:data:encdg rpbinary;:DESE 1;:*ESE 1");
    recLenNumBytes = len(recLen)
    headerLen = 1 + 1 + recLenNumBytes;
    xincr = float(scope.ask("wfmpre:xincr?"))
    ymult = float(scope.ask("wfmpre:ymult?"))
    yoff = float(scope.ask("wfmpre:yoff?"))
    yzero = float(scope.ask("wfmpre:yzero?"))
    # Determine which channels are enabled for
    if 'TD' in scopename[1]:
        if '4' in scopename[1][7]:
            channelson = range(4)
        if '2' in scopename[1][7]:
            channelson = range(2)
    else:
        if '4' in scopename[1][6]:
        if '2' in scopename[1][6]:
    # Set up the sub folder for aquired data
    filenumber = 1
    filename = scopename[1] + '_' + scopename[2] +'_'
    #Query from user of the front panel should be locked during operation
    print "Would you like to lock the front-panel during the capture session?"
    LockFP = str(raw_input())
    if (('Y' in LockFP) or ('y' in LockFP)):
        LockFP = True
        scope.write("LOCK ALL")
        capturelog.append("Front-panel locked during session? (Y/N)\n")
    else:
        LockFP= False
    # Start entering data into the capture log.
    capturelog.append("Log for captures Starting Catpures on " + ctime()+ '\n')
    capturelog.append("Tracking " + str(scopename[1]) + " for " + \
                      str(duration_time) + " minutes. \n \n")
    # Determine which channels are actually on and record to the caputre log
    for a in channelson:
        channelson[a]=int(scope.ask("SEL:CH"+ str(a+1) +"?"))
        if channelson[a] == 1:
            print 'Channel ' + str(a+1) + ' is on'
            capturelog.append('Channel ' + str(a+1) + ' is on \n')
    # Acquire screen shots for duration of time indicated
    scope.write("ACQ:STOPA SEQ")
    print "Starting Catpures on " + ctime() + '\n'
    end_time = time() + (duration_time * 60)
    while time() < end_time:
        scope.write("ACQ:STATE RUN")
        waiting = 1
        looptime = time()
        # wait here until trigger received, forced trigger occurs or time ends
        while (waiting == 1):
            waiting = int(scope.ask("ACQ:STATE?"))
            #capture the trigger time
            if waiting == 0:
                triggertime = time()
            if looptime > end_time:
                if scope.ask("TRIGGER:STATE?") == 'READY':
                    print "No trigger event for finale capture"
                    capturelog.append('\n' "No trigger event for final capture" \
                                      + '\n')
                    break
            if Forcetrig:
                if (looptime + Forcetrig_time) < time():
                    scope.write("Trigger Force")
                    print "Trigger forced at " + str(ctime(triggertime)) +'\n'
                    capturelog.append("Trigger forced at " + \
                                      str(ctime(triggertime)) +'\n')
                    while (scope.ask("TRIGGER:STATE?") != 'SAV'):
                        scope.write("Trigger Force") # allows for Averageing to be used
                    break
        # Curve data captured and saved as CSV file
        if (logmode == 1 or logmode == 2) and scope.ask("TRIGGER:STATE?") \
           != 'READY':
            for a in range(len(channelson)):
                    scope.write("DATA:SOURCE CH" + str(a+1))
                    scope.write("CURVE?")
                    datac = scope.read_raw()
                    #If FP not locked, settings may have changed
                    if LockFP == False:
                            HDelay = float(scope.ask("HORizontal:DELay:" \
                                                     +"POSITION?"))
                            HPos = float((Hscale * 5 + float(scope.ask\
                                                             ("HORIZONTAL:DELay"\
                                                              +":TIME?")))/ \
                    # Strip the header
                    datac = datac[headerLen:(int(recLen)-1)]
                    # Convert to byte values
                    datac = unpack('%sB' % len(datac),datac)
                    # Convert bytes to voltage values
                    x = []
                    y = []
                    for i in range(0,len(datac)):
                        x.append((i-(len(datac)*(HPos/100)))* xincr + HDelay)
                        y.append(((datac[i]-yoff) * ymult) + yzero)
                    #save curve data in .csv format
                    f2 = open(filepath + filename+str(filenumber)+ ' CH_' \
                              + str(a+1)+'.csv', 'w')
                    f2.write("s,Volts\n")
                        f2.write(str(x[i]) + ',' + str(y[i])+ '\n')
                    f2.close()
                    print "CH " + str(a+1) + " Waveform #" + str(filenumber) \
                          + " saved as " + filepath + filename + str(filenumber)\
                          + ' CH_' + str(a+1)+ '.csv'
                    print "     triggered at "+ ctime(triggertime) + '\n'
                    capturelog.append("CH " + str(a+1) + " Waveform #" \
                                      + str(filenumber)+ " saved as " \
                                      + ' CH_' + str(a+1)+ '.csv')
                    capturelog.append(" triggered at "+ ctime(triggertime) \
                                      + '\n')
        # start the data tansfer and save file for the Image format
        if (logmode == 0 or logmode == 2) and scope.ask("TRIGGER:STATE?") \
           != 'READY':
            # Give some time for meas values to compute if no curve data pulled.
            if logmode == 0:
                sleep(.25)
            while True:
                try:
                    scope.write("HARDCOPY START")
                    sleep(.01)
                    data = scope.read()
                    scope.ask('*OPC?')
                    break
                except VisaIOError:
                    print "***Hard copy did not terminate correctly for " \
                          + str(filenumber) + "***"
                    print"    However the image may still be OK"
                    capturelog.append("***Hard copy did not terminate correctly" \
                                      " for " + str(filenumber) + "*** \n")
                    capturelog.append("   However the image may still be OK \n")
                    break
            #Save the hardcopy data as a image file
            f = open(filepath + filename+str(filenumber)+'.'+ fileformat \
                     , 'wb')
            f.write(data)
            f.close()
            print "HardCopy #" + str(filenumber) + " saved as " \
                  + filepath + filename + str(filenumber) + '.'+ \
                  fileformat
            capturelog.append("HardCopy #" + str(filenumber) + " saved as " \
                  fileformat )
            capturelog.append(" triggered at "+ ctime(triggertime) + '\n')
        #increment the capture number
        filenumber = filenumber + 1
    #clean up after captures complete, and write the capture log to a file
    print "Completed catpures on " + ctime()+ '\n'
    if LockFP:
        scope.write("LOCK NONE")
        print "Unlocking Front-panel controlls."
    capturelog.append('\n' + "Completed all on " + ctime()+ '\n')
    # Save capture log to file
    caplogfile = filepath + "Capture_Log_" + str(int(time())) + ".txt"
    print "Saving Capture Log to " + caplogfile
    f3 = open(caplogfile, 'w')
    for a in range(len(capturelog)):
        f3.write(capturelog[a])
    f3.write("Initial scope setup:\n")
    f3.write(Scopesetup)
    f3.close()
    scope.write(Scopesetup)
    print "Done"
This is my first attempt writing such a program, and my first time using Python. Any recommendations would be appreciated.