Keithley 220 Voltage limit?
appleSwell
Post by appleSwell » June 22nd, 2018, 12:15 am
Hi everyone,
I set the V-Limit of Keithley 220 as 100V, but it cannot go past the current at which the voltage hits around 25V. Is there any other setup that might put a cap in voltage before the V-Limit we set on the front panel of 220? I'm running it via LabVIEW programs.
brian.d.smith
Re: Keithley 220 Voltage limit?
Post by brian.d.smith » July 5th, 2018, 6:31 am
Maybe you are shorting and/or loading the GUARD? The datasheet says that total (output + guard) current cannot exceed 105mA.
Make sure that you use an adapter that drops guard and does not short guard to LO/SHIELD.
You must use an appropriate triax adapter at the rear of the instrument. Using a BNC adapter will likely ruin the triax jack on the rear of the instrument.
