Voltage drift with time in 2182A

Post by spkeashwar » December 29th, 2018, 11:38 am

Hello,

I am using 6221 to source a current(20uA) and 2182A to measure the voltage(0.432mV) across a DUT. However I see that the voltage builds up over time and it takes a few minutes for it to reach a stable value. The DUT is fairly metallic as can be seen from its resistance value. I don't understand what is the reason for this voltage build up over time. It gets worse in devices with higher impedance where it takes longer time to stabilize.
V build.png
V build.png (13.83 KiB) Viewed 2305 times
Thank you,
Pranava.

Andrea C
Keithley Applications
Keithley Applications
Posts: 1287
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: Voltage drift with time in 2182A

Post by Andrea C » January 1st, 2019, 7:24 am

Assuming instruments are warmed up and room environment stable.
It looks like beautiful first order response.
At 20uA and low ohms, the capacitance is massive to have this time constant.
What do your connections look like?
Can you try different levels of current.
Try sourcing negative current. Same RC outcome?