it takes me 300milliseconds to read pk2pk value.can it be faster?
kenash
Country: Russian Federation
Post by kenash » September 10th, 2021, 1:49 am
it takes me 300milliseconds to read pk2pk value from tds2024c. can it be faster?say,30milliseconds?
Am i doing it right? below is my c++ code
        status = viWrite(vi, (ViBuf) "MEASUREMENT:IMMED:VALUE?", 24, &retCnt);
	if (status < VI_SUCCESS) {
		HandleError();
		return 1;
	t = std::chrono::system_clock::now();
	// Clear the buffer and read the response
	memset(buffer, 0, sizeof(buffer));
	status = viRead(vi, (ViBuf)buffer, sizeof(buffer), &retCnt);
		HandleError();
		return 1;
	return 0;
Dave W
Re: it takes me 300milliseconds to read pk2pk value.can it be faster?
Post by Dave W » September 23rd, 2021, 8:34 am
Hi Kenash,
That's not too surprising that it would take that long on a TDS2024C. When you do an immediate measurement ("MEASURMENT:IMMED") the instrument will calculate the value upon receiving that command and does not use a cached value (for example any value seen on screen). The TDS2000C series doesn't have the beefiest processor in the world so it can take a little time to recalc that measurement, especially if the scope is still in run mode and acquiring new data.
To speed it up you can try stopping the acquisitions if you haven't already. You can also try just reading back on-screen measurement values, although do note that there is a slight delay from the point of an acquisition completing and the on-screen measurement updating so this works best if the scope is just in free-run acquisition mode rather than in single sequence acq. mode.