Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

2510 resistance measurements incorrect

Post Reply
cballester
Posts: 1
Joined: March 22nd, 2012, 8:28 am
Country: United States

2510 resistance measurements incorrect

Post by cballester » March 22nd, 2012, 8:37 am

The issue we are having is that sometimes the units report very low or even negative resistance. If you reboot the instrument, the problem goes away. It does not matter what device and what instrument we use, they all have this issue.

This happens on at least 7 instruments that we have, and they have firmware revisions ranging from A03 to A08.

Dale C
Keithley Applications
Keithley Applications
Posts: 2847
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: 2510 resistance measurements incorrect

Post by Dale C » March 22nd, 2012, 11:14 am

What are you trying to measure?
If you are measuring resistance of an RTD or Thermistor?
If so, what setpoint? 1kohms, 10kohms?, etc.